Ventilator alarms and troubleshooting

Sort alarms into high pressure, low pressure or volume, and apnea, and use DOPE when a ventilated patient suddenly gets worse.

6 practice questions3 min read5 sourcesUpdated Oct 30, 2026

  • First moveLook at the patientNot the screen
  • Unstable or unsureDisconnect and bagWith 100% O2 while you find the cause
  • Cuff pressure20–30 cm H2OCheck with a manometer

An alarm means the ventilator measured something outside the limits you set. It does not say why. Your job is to look at the patient first, then match the alarm to a cause.

Look at the patient

Check color, chest rise, work of breathing, SpO2, heart rate, blood pressure, and the ETCO2 waveform before you touch a knob. A patient who is getting worse while the alarm sounds needs a different response than a patient who coughed.

High-pressure alarm

Airway pressure went above the limit during inspiration. Something is raising resistance or lowering compliance.

  • Resistance (PIP up, Pplat unchanged): secretions or a mucus plug, patient biting the tube, kinked tube or circuit, water in the circuit, bronchospasm.
  • Compliance (PIP and Pplat both up): pneumothorax, pulmonary edema, worsening ARDS, tube in the right mainstem bronchus, abdominal distension.
  • Patient effort: coughing, or fighting the ventilator (asynchrony).

A sawtooth pattern on the flow or volume waveform and coarse breath sounds say secretions. Suction. See ventilator graphics for the PIP versus Pplat check.

Low-pressure or low-volume alarm

Pressure or exhaled volume came in below the limit, so gas is going somewhere other than the lungs.

  • Disconnection at the wye, humidifier, or tube connector.
  • Circuit leak: loose connection, cracked tubing, leaking exhalation valve or humidifier.
  • Cuff leak: low cuff pressure, a cuff that has ruptured, or a tube that has moved up so the cuff sits at the cords.
  • Accidental extubation: the tube is out or the cuff is above the cords.
  • Chest tube or bronchopleural fistula air leak.

Check the cuff with a manometer and aim for 20–30 cm H2O. Feeling the pilot balloon is inaccurate.

Apnea alarm

The ventilator did not detect a patient-triggered breath within the set time. Causes include sedation or neurologic depression, weak effort, a trigger set too insensitive, or a disconnect. Make sure the backup ventilation is on and the patient is actually being ventilated.

DOPE

When a ventilated patient suddenly gets worse, work through DOPE:

Letter Cause Check
D Displacement of the tube (out, or in a mainstem bronchus) Waveform capnography, depth at the teeth, breath sounds on both sides
O Obstruction of the tube or circuit Pass a suction catheter, look for biting or kinks
P Pneumothorax Breath sounds, trachea position, chest X-ray or ultrasound
E Equipment failure Disconnect and bag. If the bag is easy to squeeze and the chest rises, the fault is the ventilator or circuit

Some versions add S for stacked breaths (auto-PEEP), which disconnecting also relieves. For a tracheostomy, see tracheostomy care.

Study example

The high-pressure alarm sounds and SpO2 drops to 84%. The patient looks anxious and the ETCO2 waveform is still present. You disconnect and bag; it is hard to squeeze and the left chest rises less. Breath sounds are absent on the left. The tube was at 23 cm at the teeth and is now at 27, so it is not a ventilator fault but a tube that has moved. Pull it back, recheck breath sounds and the ETCO2, and get a chest X-ray.

See ventilator graphics, endotracheal intubation, airway suctioning, and pneumothorax.

Sources

  1. WikEM. Deterioration after intubation (DOPES, disconnect and bag, ETCO2 and tube depth).
  2. Goodfellow LT, et al. AARC Clinical Practice Guideline: Patient-ventilator assessment. Respir Care. 2024;69(8):1042-1054.
  3. Blakeman TC, et al. AARC Clinical Practice Guidelines: Artificial airway suctioning. Respir Care. 2022;67(2):258-271.
  4. Emrath E. The basics of ventilator waveforms. Curr Pediatr Rep. 2021.
  5. Merck Manual Professional Edition. Tracheal intubation (confirming tube placement).